Datasheet ADuM6200, ADuM6201, ADuM6202 (Analog Devices) - 10
DescriptionDual-Channel, 5 kV Isolators with Integrated DC-to-DC Converter
Pages / Page28 / 10 — ADuM6200/ADuM6201/ADuM6202. Data Sheet. ABSOLUTE MAXIMUM RATINGS. Table …
RevisionC
File Format / SizePDF / 576 Kb
ADuM6200/ADuM6201/ADuM6202. Data Sheet. ABSOLUTE MAXIMUM RATINGS. Table 19. Parameter. Rating. ESD CAUTION
ADuM6200/ADuM6201/ADuM6202 Data Sheet ABSOLUTE MAXIMUM RATINGS Table 19 Parameter Rating ESD CAUTION
link to page 18 link to page 9 link to page 23 link to page 10 link to page 10 link to page 10 link to page 10 link to page 10 link to page 10
TA = 25°C, unless otherwise noted. Stresses above those listed under Absolute Maximum Ratings may cause permanent damage to the device. This is a stress
Table 19.
rating only; functional operation of the device at these or any
Parameter Rating
other conditions above those indicated in the operational Storage Temperature (T ) −55°C to +150°C ST section of this specification is not implied. Exposure to absolute Ambient Operating Temperature (T ) −40°C to +105°C A maximum rating conditions for extended periods may affect Supply Voltages (V , V )1 −0.5 V to +7.0 V DD1 ISO device reliability. Input Voltage (V , V , V , V , V , −0.5 V to V + 0.5 V IA IB E1 E2 SEL DDI RC , RC )1, 2 IN SEL Output Voltage (V , V )1, 2 −0.5 V to V + 0.5 V
ESD CAUTION
OA OB DDO Average Output Current per Pin3 −10 mA to +10 mA Common-Mode Transients4 −100 kV/µs to +100 kV/µs 1 Each voltage is relative to its respective ground. 2 VDDI and VDDO refer to the supply voltages on the input and output sides of a given channel, respectively. See the PCB Layout section. 3 See Figure 5 for maximum rated current values for various temperatures. 4 Refers to common-mode transients across the isolation barrier. Common- mode transients exceeding the absolute maximum ratings may cause latch-up or permanent damage.
Table 20. Maximum Continuous Working Voltage1 Parameter Max Unit Applicable Certification
AC Voltage, Bipolar Waveform 424 V peak All certifications, 50-year operation AC Voltage, Unipolar Waveform Basic Insulation 600 V peak Reinforced Insulation 565 V peak Working voltage per IEC 60950-1 DC Voltage Basic Insulation 600 V peak Reinforced Insulation 565 V peak Working voltage per IEC 60950-1 1 Refers to the continuous voltage magnitude imposed across the isolation barrier. See the Insulation Lifetime section for more information. Rev. C | Page 10 of 28 Document Outline Features Applications General Description Functional Block Diagrams Table of Contents Revision History Specifications Electrical Characteristics—5 V Primary Input Supply/5 V Secondary Isolated Supply Electrical Characteristics—3.3 V Primary Input Supply/3.3 V Secondary Isolated Supply Electrical Characteristics—5 V Primary Input Supply/3.3 V Secondary Isolated Supply Package Characteristics Regulatory Information Insulation and Safety-Related Specifications Insulation Characteristics IEC 60747-5-2 (VDE 0884 Part 2):2003-01 and DIN V VDE V 0884-10 (VDE V 0884-10):2006-12 Thermal Derating Curve Recommended Operating Conditions Absolute Maximum Ratings ESD Caution Pin Configurations and Function Descriptions Truth Table Typical Performance Characteristics Terminology Applications Information PCB Layout Start-Up Behavior EMI Considerations Propagation Delay Parameters DC Correctness and Magnetic Field Immunity Power Consumption Current-Limit and Thermal Overload Protection Power Considerations Thermal Analysis Increasing Available Power Insulation Lifetime Outline Dimensions Ordering Guide